Contributions:NihonKohdenSource
Nihon Kohden
The Nihon Kohden JE-120A amplifier is a clinical monitoring system capable of acquiring EEG or ECoG signals from up to 256 channels. Paired with the QI-123A LAN converter box, this system can provide a research data stream in addition to the clinical data stream. The research data stream can be sampled at up to 2kHz for 256 channels, 5kHz for 128 channels or 10kHz for 64 channels.
http://www.bci2000.org/svn/trunk/src/private/SignalSource/NihonKohdenSource
Griffin Milsap (griffin.milsap@gmail.com) Johns Hopkins University
Peter Brunner (brunner@neurotechcenter.org) Washington University in St. Louis
Markus Adamek (adamek@neurotechcenter.org) Washington University in St. Louis
Alexander Belsten (belsten@neurotechcenter.org) Washington University in St. Louis
  1. Version 1.0 (r4913): November 2015 - Initial release.
  2. Version 1.1 (r6245): March 2021 - Updated to support x64.
  • Initial development: 4913
  • Tested under: 6245
  • Known to compile under: 6245
This setup requires the following components:
  • JE-120A amplifier:
  • QI-123A LAN converter:
The amplifier is a clinical monitoring system capable of acquiring EEG or ECoG signals from up to 256 channels. Paired with the QI-123A LAN converter box, this system can provide a research data stream in addition to the clinical data stream. The research data stream can be sampled at up to 2kHz for 256 channels, 5kHz for 128 channels or 10kHz for 64 channels.
  • Input range: ±3.2 mV or ±12.8 mV (selectable)
  • Effective resolution: 16 bit (97.65 nV/LSB) or 16 bit (390.62 nV/LSB) (selectable)
  • Input impedance: 200 MΩ
  • Sampling Rates: 100Hz, 200Hz, 500Hz, 1000Hz, 2000Hz, 5000Hz, 10000Hz
Clinical Setup
To enable the research data stream, enable "Use QI-124A ports" under the JE-120 settings.
alt text
Installation
This source module is located in a private directory of BCI2000. To build the source, you must have the necessary credentials, enable BUILD_PRIVATE in CMake, and modify the CMakeLists.txt in the private and SignalSource directories to build to NihonKohdenSource.
  • SourceCh total number of channels to be acquired from the source module or auto for all channels enabled in the Neuroworkbench
  • SampleBlockSize number of samples acquired in each block or auto for default setting (1/10th of sampling rate)
  • SampleRate determines the rate at which the device samples data.
  • ChannelNames is a convenience parameter. Name channels here and they can be referenced by these names later.
  • SourceChOffset auto for default settings (0 for each channel).
  • SourceChGain auto for default settings (1 for each channel).
  • SourceChList list of channels that should be acquired or auto for all channels enabled in the Neuroworkbench
  • SourceBufferSize
Valid Operating Modes
The NihonKohdenSource only accepts specific sampling rates. The number of channels acquired is limited at sampling rates of 5000 Hz and above. The following table shows all valid combinations of sample rate and maximum number of channels.
SamplingRate Max SourceCh Aliasing Filter Cut-Off
[samples/sec] [channels] [Hz]
100 256 30 or 35
200 256 60 or 70
500 256 120
1000 256 300
2000 256 600
5000 128 1200
10000 64 3000
  • DC01 to DC16 DC Input data
    • Digital Range: 0 to 65535
    • Analog Range: 0xFFFF(Max)-0x8000(0V)-0x0000(Min). (Signed.)
    • 1LSB: 366.3uV
  • SpO2
    • Digital Range: 0 to 100
    • 0 means 0%, 100 means 100%
  • EtCO2
    • 0 means 0mmHg, 100 means 100mmHg
  • BP1 to BP2
    • not implemented
  • DigitalMark1 to DigitalMark2
    • not implemented
DC Input From Other Devices
Analog data from a g.TRIGbox and parallel port can be obtained though the JE120A's DC inputs. A parallel port Y-splitter and a custom cable is necessary. Connection of the custom cables and Y-splitter is outlined in the image below.
alt text
The outputs from the g.TRIGbox and parallel port to the DC input of the JE120A are defined below. Note that the g.TRIGbox trigger outputs are the TTL pins. See Contributions:ParallelPortFilter for information on how to set the outputs of a parallel port from BCI2000.
JE120A g.TRIGbox Parallel Port
Pin No. Signal Pin No. Signal Pin No. Signal
1 DC01 2 Output A - -
2 DC02 3 Output B - -
3 DC03 4 Output C - -
4 DC04 5 Output D - -
5 DC05 8 Encoded Output - -
6 DC06 - - - -
7 DC07 - - - -
8 DC08 - - - -
9 DC09 - - 2 Data0
10 DC10 - - 3 Data1
11 DC11 - - 4 Data2
12 DC12 - - 5 Data3
13 DC13 - - 6 Data4
14 DC14 - - 7 Data5
15 DC15 - - 8 Data6
16 DC16 - - 9 Data7
24 GND 1 GND 24 GND
Certification Results
The Nihon Kohden certification was performed to give a better understanding of the restraints that must be placed on the system when using this device. Certification success criteria was determined internally and without rigor, and reflects our experience with psychophysical experiments and the performance we expect to need for the system. More information about certification testing can be found on the certification wiki page.
PC Specifications
Certification Tests Performed
Certification testing varied number of channels (4, 8, 16) task (P3Spell (7x7 grid), P3Spell (1x1 grid), CursorTask, CursorTaskLow, and StimulusPresentation), and tested audio and video latency, for a total of 15 performed tasks.
Success Criteria and Overall Results
Name Actual Latency Success Criteria Result
Timestamp Latency Mean 2.81 µs 1 ms passed
Timestamp Latency STD 2.75 ms 5 ms passed
Processing Latency Mean 35 ms 20 ms failed
Processing Latency STD 164 ms 10 ms failed
Video Latency Mean 66.9 ms 65 ms failed
Video Latency STD 6.31 ms 20 ms passed
Audio Latency Mean 76.2 ms 65 ms failed
Audio Latency STD 3.94 ms 20 ms passed